6 D4 ^* z( t4 ?8 S+ R | Measurement of Conductivity
, J, m6 y" R# P3 P | 7.3: b1 j, K. z8 T0 d. U! `5 W
| 60112(ed.4)1 G( P0 q, V/ M
| * Q8 O0 x( @' g1 @3 }3 B) Y h6 e/ I
Question:( T6 _* T7 L# s3 ^% n
Clause 7.3 requires that the conductivity of the tracking test solution “be measured with alternating voltage at a frequency in the range 1 kHz to 2 kHz. The procedure is described in IEC 60589”. The procedure in IEC 60589 utilises conical flasks with reflux condensors, platinised platinum electrodes, a conductivity cell and a resistance bridge. This appears to be a chemical standard rather than an appropriate electrical test laboratory measurement technique.9 ^1 {8 c X! x" ^4 X- A
Are CBTL:9 C% V- C. q: j: f
- (a) expected to use this method of measurement for solution conductivity; or
* `5 c9 H# j& f4 ? }: h- (b) can a suitable conductivity meter with a measurement frequency in the range 1 kHz to 2 kHz be used; or# ~( T! F! J% r0 u, P
- (c) can any suitable conductivity meter (different measurement range) be used.9 `& V; A8 B, J
Decision:* U' R. B$ k0 [4 i S' O, R
The Group decided that the “use of any appropriate conductivity meter with sufficient accuracy and uncertainty including the probe” is the only practical one - item (c) above.1 H7 J2 w7 d% d; f: k