请教一下, 在1500V,或3000V耐压测试中,为什么漏电流在耐压仪里设定是10mA? 在IEC/UL60950里有规定吗?7 v+ A8 b2 v: h0 N3 K
是否是这一节: 6.1 q$ g2 [4 y8 t1 `% \7 B0 R
For equipment powered from an AC MAINS SUPPLY, the test is performed with a voltage equal to% W1 F5 b: p8 ^# X* P
the RATED VOLTAGE of the equipment or to the upper voltage of the RATED VOLTAGE RANGE. For9 M T6 o! J" n0 X. @
equipment powered from a DC MAINS SUPPLY, the test is performed with a voltage equal to the
+ \8 M' X) _) @( f' ?. hhighest nominal voltage of the AC MAINS SUPPLY in the region where the equipment is to be# x1 `8 {: Q( Y! P7 H
used, for example, 230 V for Europe or 120 V for North America.
j6 s7 G5 s: k4 C! zThe current flowing in the test circuit of Figure 6A shall not exceed 10 mA.
/ L9 ^! V1 P( U% V+ d但我认为这一章是对通迅网络而言的. 不是对整机和变压器而言的. |
|