| DSH 546
& r, J$ h7 B5 L
3 @* e" q- F f# }* n | Relative humidity
. W7 G5 V2 P4 v | 10.2 (IEC 60065/2001) 15.3 (IEC 60335-1/2001) 4.10 (60601-1/1988 +A2/1995) 6.8.2 (IEC 61010-1/2001)- m! y1 |% m! [: y
| 60335-1(ed.4) & 61010-1(ed.2) & 60601-1(ed.2);am1;am2 & 60065(ed.7)7 O8 `7 @, S h7 p! K
|
$ o: n8 R+ N2 V: U+ {Question:( O% m8 V3 ?) E+ y8 U; K+ K* } |
Various IEC standards require that specific tests be performed in a humidity chamber. The& q/ @) z% o! c6 |7 b' _
specification of the relative humidity is in several formats. How should the tolerance on the
: q* |7 ]/ M" I+ z7 |' {relative humidity specification be interpreted?5 U* w/ S' I% Q y c! p( v
Decision:
. ~ H, c, v# d( H' C. x. R(1) Unless clearly noted otherwise, the tolerance specified for humidity is considered to
) E0 p2 g, h- D$ A# r8 `% Fbe in percent relative humidity, % RH.
% T& y1 H4 i$ F(2) It is intended that the set point of the humidity chamber controller be set within the' @4 n+ x: c) F
specified range and that the humidity remains within the specified range.
$ Z9 a" }# k" x. K) d# S(3) Confirmation that humidity is within the specified range shall be made by a humidity/ M) f! V& l* u, a* I4 p
measuring instrument according to CTL DSH 251A.4 m$ s4 W# C! l8 {' B' p( M