Standard(s)- (year and edition):
) y" a: ?: v& i/ L; o3 bIEC 60601-1:1988 Ed.2 Am1+Am2' x% h& I! L! H- V: w5 @
Sub clause(s): 16, 20, 19.2
: S: X- z4 @* D2 [2 vSheet n°: DSH-410/ ~0 L' `6 f% ]: {3 X, b
Subject: Use of dc supply
4 K, h+ H+ J0 o! \Key words: dc supply, enclosure( q, [" |. J: `% I1 j! P
Confirmed by CTL at its 39th meeting, in Cologne
( g( S9 M! w! n) r: n* J0 M, G5 _Question:
" ?- T% @3 B6 i9 mSome EQUIPMENT is designed for use in ambulances and operated from the vehicle d.c.
/ }1 w' a. x7 P" F9 k7 Qsupply.with the negative side of the supply connected to the enclosure. Is this acceptable?
/ [1 X7 C/ i* y6 |Decision:
! Y! }7 \6 H& k+ e+ V O- ~Connection of the negative side of the d.c. mains to the ENCLOSURE should not be permitted.
4 {' w/ H6 C1 K2 p/ AThe insulation between all poles of the supply and the ENCLOSURE must comply with 20.1 A-a1' i9 J* r& Q9 V _7 |
and A-a2.) Z+ R9 q" Y5 @% _: B
Note: Vehicle chassis is considered to be protective earth.$ `# l+ i, G4 c6 f2 a& W5 _9 P: W