' w7 w" |+ O1 b7 I+ E7 X7 jStandard(s)- (year and edition):7 J5 z5 u$ K5 W: w1 `$ l
IEC 60601-1:1988 Ed.2 Am1+Am26 R5 @5 @. a& {, B/ N1 Z% c- R
Sub clause(s): 57.9 57.9.4; u0 [0 W# a6 I7 w5 a' x9 V# S
Sheet n°: DSH-426
+ A \+ B4 H9 m, uSubject: Triple insulated windings
) k; f* S7 x. p" H! N/ s' P0 xKey words: Triple, insulated, winding, transformer) K; K: ?# g. e: _. h
Confirmed by CTL at its 39th meeting, in Cologne6 i2 T$ h* ^5 B. `
Question:( P2 q& W& S5 A9 a
Many manufacturers are using a triple insulated winding wire on a transformer, where
" k6 m; x- u# V4 C& s2 |" q1 Q6 ?DOUBLE, or REINFORCED INSULATION is required between the primary and the secondary" K, v f% Z4 [9 p
windings. Is this permitted?) l; e& S G a/ c
Decision:1 |# C: x4 l4 {: c5 b5 V
This is allowed, but precautions should be taken concerning possible mechanical damages to1 |* X9 @ o, O; |0 C3 \3 [9 N
the wires. Other requirements of 57.9 shall be applied.