EMC中传导骚扰检测一定要在屏蔽室内进行吗?
请教大家:EMC中传导骚扰检测一定要在屏蔽室内进行吗? In most of EMC standards, the interference voltage test (Line Conduct Test) is not particularly required to have the test in a shielded room. However, to eliminate the affects coming from the background noises, almost all laboratories would prefer getting this test in the shielded room. 因骚扰电压是属于线上干扰,而非空间干扰,标准中未规定骚扰电压一定要在屏蔽室内进行
但需要有良好的测试环境:
1.测试电源需稳定且无干扰
2.在试验台下方水平位置和靠墙方垂直位置应有两块足够大的金属接地平板
但骚扰功率这一项目是属于空间传播的干扰,因此需在屏蔽室内进行 一定要在屏蔽室内做,一共包括两项 LISN 和ISN,EUT 要距离屏蔽室0.5米处,距离地面0.8米要有至少大于实验桌子0.5米长宽的接地平板。