DSH 316 Mains Transformers Overload test
http://decisions.iecee.org/icons/ecblank.gifDSH 316
Mains Transformers Overload test
14.7.2
61010-1(ed.1);am1;am2
Standard:
IEC 61010-
1:1990+A1:1992+A”:1995
Sub clause:
14.7.2
Sheet n. 316
Page 1(1)
Subject:
Mains transformers
Overload test
Key words:
- Transformer
- Overload
- Temperature
Decision taken by
ETF3 and confirmed
by CTL at its 38th
meeting, in Toronto
Question:
Should this test be performed starting from cold condition or from operating temperature?
Decision:
Tests from both temperatures are valid and should therefore, be performed if there is
likelihood that limits will be exceeded.
Explanation
All working temperatures are addressed by this requirement. Laboratories should analyse the
operation of the device to determine whether more than one test is needed.