DSH 724 Measurement of Conductivity
http://decisions.iecee.org/icons/ecblank.gif DSH 724
Measurement of Conductivity
7.3
60112(ed.4)
Question:
Clause 7.3 requires that the conductivity of the tracking test solution “be measured with alternating voltage at a frequency in the range 1 kHz to 2 kHz. The procedure is described in IEC 60589”. The procedure in IEC 60589 utilises conical flasks with reflux condensors, platinised platinum electrodes, a conductivity cell and a resistance bridge. This appears to be a chemical standard rather than an appropriate electrical test laboratory measurement technique.
Are CBTL:
- (a) expected to use this method of measurement for solution conductivity; or
- (b) can a suitable conductivity meter with a measurement frequency in the range 1 kHz to 2 kHz be used; or
- (c) can any suitable conductivity meter (different measurement range) be used.
Decision:
The Group decided that the “use of any appropriate conductivity meter with sufficient accuracy and uncertainty including the probe” is the only practical one - item (c) above.
页:
[1]